Question:

Arrange the instruments in increasing order of their sensitivity:
A. GC-ECD
B. TLC
C. HPLC-PDA
D. GC-TCD
E. HPLC-RI

Show Hint

As a rule of thumb for sensitivity: Universal detectors (RI, TCD) are often less sensitive than selective detectors (PDA, ECD). The \textbf{Electron Capture Detector (ECD)} is famously sensitive for specific classes of compounds like pesticides and PCBs.
Updated On: Sep 23, 2025
  • A, B, C, D, E
  • B, E, C, A, D
  • C, E, A, D, B
  • E, B, D, C, A
Hide Solution
collegedunia
Verified By Collegedunia

The Correct Option is D

Solution and Explanation

Sensitivity refers to the lowest concentration of an analyte that an instrument can reliably detect. We need to order these techniques from least sensitive to most sensitive.
Step 1: Order the techniques from least to most sensitive.
- E. HPLC-RI (Refractive Index Detector): Generally considered one of the least sensitive detectors in chromatography. It is universal but has poor detection limits.
- B. TLC (Thin-Layer Chromatography): This is a visual, manual technique. Its sensitivity is relatively low, typically in the microgram range. It is generally more sensitive than RI but less sensitive than most other instrumental detectors.
- D. GC-TCD (Thermal Conductivity Detector): A universal but moderately sensitive detector for GC. Less sensitive than FID or ECD.
- C. HPLC-PDA (Photo Diode-Array Detector): A UV-Vis detector that offers good sensitivity for compounds that absorb UV light. Generally more sensitive than TCD.
- A. GC-ECD (Electron Capture Detector): This is an extremely sensitive and selective detector for compounds containing electronegative atoms (like halogens in pesticides). It is one of the most sensitive detectors available.

Step 2: Form the sequence.
The order from least sensitive to most sensitive is: HPLC-RI<TLC<GC-TCD<HPLC-PDA<GC-ECD.
The corresponding letter sequence is E \(\rightarrow\) B \(\rightarrow\) D \(\rightarrow\) C \(\rightarrow\) A.