Question:

Which of the following statements related to the operating principle of pulse oximetry are CORRECT?

Show Hint

Pulse oximeters rely on the different absorption spectra of oxyhemoglobin and deoxyhemoglobin at red and infrared wavelengths.
Updated On: Dec 4, 2025
  • Pulse oximeter can non-invasively determine arterial oxygen saturation (SpO$_2$) by analyzing the light transmitted through the skin during the systolic phase of the blood flow through the tissue
  • In a pulse oximeter, isosbestic wavelength is the wavelength at which Hb and HbO$_2$ have same optical absorbance
  • Pulse oximeter can accurately determine the SpO$_2$ of blood by computing the ratio of absorbances at 660 nm and 905 nm wavelengths
  • Pulse oximeter can accurately determine the SpO$_2$ of blood by computing the ratio of absorbances at 850 nm and 950 nm wavelengths
Hide Solution
collegedunia
Verified By Collegedunia

The Correct Option is A, B, C

Solution and Explanation

(A) Pulse oximeters use transmitted light during systole to measure arterial oxygen saturation, so this statement is true.
(B) An isosbestic point is indeed a wavelength where Hb and HbO$_2$ absorb light equally — correct.
(C) Standard pulse oximeters use red (660 nm) and IR (905–940 nm), so this statement is correct.
(D) The pair 850–950 nm is not used for SpO$_2$ measurement — incorrect.